Posted - May 04 2010 : 12:29:32 PM
|
Hello all,
I'm double-checking on a device safety profile for a patient we are hoping to scan on BIAC3. The device is noted by the manufacturer as being MR Conditional with the following limitations:
1. 3.0 or less Tesla 2. Spatial Magnetic Gradient Fields <= 1000 Gauss/cm 3. Whole body SAR 4 W/kg for <= 15 min.
Numbers #1 and #3 are easy to check on, it's number #2 I'm uncertain about. Does anyone know what the spatial magnetic gradient field strength is for BIAC3? Sequence dependent?
Thanks, Jeff |
|